long run to test the appearance of high frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
today, long run directly at full power (33% amplifier ratio) => 98kW in the FPC at the very begining of the run.
(P=0.05 / I=0.0005 / D=0.6)
10-40 min : almost misalignment effect (CEP did not change so much)
I increased 3 times the amplifier ratio to 34%, 35%, 36% to compensate the misalignment and keep the power almost constant.
then I did a full tuning (CEP + alignment)
45-110 min : misalignment effect after ~30min of warming up of the FPC => much more stable.
I corrected twice (@ 85min) the alignment to compensate a bit the power loss
globally, the FPC seems stable => all the lock losses come from the 20Hz noise and are recovered very quicly by the locking.
I never saw the high frequency noise which can produce long lock losses.
maybe it's time to add the gain x3 on the laser PZT channel to get some room on the 20Hz noise compensation.
| Ronic Chiche wrote: |
|
Today with Daniele, we did 6 long runs at different power (23kW, 46kW, 66kW, 73kW, 92kW, 92kW)
All the lock loss in between these several runs are due to FPC locking parameters change.
most of the few lock losses during the stable power duration, are due to 20Hz oscillation noise or because we forgot to center properly the PZT in its range (operator faults).
the 4 first runs (23kW, 46kW, 66kW, 73kW) are using the PID : (P=0.1 / I=0.0015 / D=1.5) without the additionnal optical attenuator placed after the diffuser.
the 5th run (92kW) is using the PID : (P=0.1 / I=0.0015 / D=1.5) with the additionnal optical attenuator placed after the diffuser.
the 6th run (92kW) is using the PID : (P=0.05 / I=0.0005 / D=0.6 and obviously a different diffuser position) with the additionnal optical attenuator placed after the diffuser.
surprisingly, we never saw any high frequency noise during the day !
to be noticed : the electron machine was OFF / the day was sunny without wind / almost nobody was working in the bunker.
|
|
laser PZT amplifier, posted by Ronic Chiche at ThomX igloo about detectors and electronics
|
this afternoon, I added the Gain = 2.8 amplifier at the output of the Laselock to drive the laser PZT.
the power supply of the board is disymetric to address the correct maximal dynamic range of the amplifier (~ 30V) and voltage drop (~ 2V) related to the power supply voltage.
so, the power supply is set to -2V / 32V which gives roughly 0-28V of dynamic range at the ouput for 0-10V at the input.
despite the additionnal noise added by the amplifier, we are able to lock easily, and reach ~ 93kW in the FPC for 33% amplifier ratio.
obviously, as the dynamic range of this PZT has been multiplied by ~3, it is much more comfortable to operate the motors.
but the overall stability seems a bit degraded... to be checked, as finding the right PID + fast loop gain is not easy.
=> to be done tomorow. |
New recipies with additionnal attenuation before the diffuser and new x3 amplifier on laser PZT, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
With optical attenuation + x3 amplifier on laser PZT channel + feedback tuning
Temporary recipie for ~ 92kW power in the FP-cavity:
(in the previous scheme, without x3 laser PZT amp, a good lock was achieved for a PDH signal power noise ~ 70 mV rms)
- Alplhanov amplifier ratio : 34%
- Axis 18 position : +0016956 steps
- PID A (for laser/FPC lock) :
input : input a
P = 0.04
I = 0.0005
D = 0.48
Sign : positive
Sampling : fast
filter : off
criterion : input d
upper th. : 10V
lower th. : 0.03V
speed : 100V/s
relock mode : none
off mode : reset
range : 50%
offset : 5V
- PID B (for RF/FPC lock) :
input : input b
P = 1
I = 0.00005
D = 1
Sign : positive
Sampling : mid
filter : off
criterion : input c
upper th. : +0.4V
lower th. : -0.45V
speed : 1V/s
relock mode : none
off mode : reset
range : 50%
offset : 5V
| Ronic Chiche wrote: |
|
For ~ 92kW power in the FP-cavity:
- Alplhanov amplifier ratio : 33%
- Axis 18 position : +0003768 steps
- PID A (for laser/FPC lock) :
input : input a
P = 0.1
I = 0.0015
D = 1.5
Sign : positive
Sampling : fast
filter : off
criterion : input d
upper th. : 10V
lower th. : 0.05V
speed : 100V/s
relock mode : none
off mode : hold
range : 50%
offset : 5V
- PID B (for RF/FPC lock) :
input : input b
P = 5
I = 0.0001
D = 1
Sign : negative
Sampling : mid
filter : off
criterion : input c
upper th. : 0.45V
lower th. : -0.45V
speed : 1V/s
relock mode : none
off mode : hold
range : 20%
offset : 5V
|
|
Tuning of the slow feeback loop analog BW + fast feedback loop gain + PID, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
Today with Daniele, we tuned the potentiometer after the Laselock which drives the analog BW of the PID on the laser PZT,
and we tuned also the fast loop gain (with the rotary potentiometer dedicated to it) => the PID parameters have changed.
so, the recipies has changed => to be updated
the global observation with the x3 amplifier on the laser PZT channel, is the system is more noisy.
the maximum power for 33% amp ratio is now ~93kW instead of 97-98kW.
we tried this afternoon to do a long run with the RF and the FPC loops activated, but it seams that a lot of high frequency noise is present.
is it coming from the new parameters feedback setup or because of the electron machine which is ON ? |
Alignment correction during the summer break, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
I restarted the FPC system this morning.
after some classic alignement procedure (some LEFT steps on the cavity injection motors X & Y) and CEP tuning,
I got 92kW for 33% amplifier ratio.
the cavity was not particularely misaligned...
then I did a long run at 90kW with both feedbacks ON without any problem. |
Power shut down on August 29th, posted by Ronic Chiche at ThomX igloo about detectors and electronics
|
This morning, all the equipments have been switched off due to a power shut down tomorrow morning.
Only the Onefive oscillator is still ON, connected to the Uninterruptable Power Supply (UPS). |
Power shut down on August 29th, posted by Ronic Chiche at ThomX igloo about detectors and electronics
|
after alignment, CEP adjustment and optimizing the PDH phase, I got back ~92kW in the FPC for 33% amplifier ratio.
| Ronic Chiche wrote: |
|
This morning, all the equipments have been switched off due to a power shut down tomorrow morning.
Only the Onefive oscillator is still ON, connected to the Uninterruptable Power Supply (UPS).
|
|
polarization measurement of the FP cavity beam, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
today, with Daniele, we did a simple polarization measurement in transmission of the FP cavity beam.
1) we locked the FP cavity and measured 82.5kw (for 33% amplifier ratio) with the power-meter moved a little bit further (this is maybe the reason why the power is a little bit lower than usually measured).
2) then, we installed a high power PBS CCM1-PBS25-1064-HP(/M) with a transmission of 89% of the P-polarization @ 1030nm
We measured 66.5kW in transmission of the cube.
3) then we installed a half waveplate to find the optimum angle => 77kW measured at an angle of 342°
(almost no power has been observed in the vertical polarization state of the PBS).
Then we rotated the waveplate to find the previous measurement => 66.5kW at an angle of 348°
(we checked that adding the waveplate in the path almost does not change the measured power)
Conclusions :
- the FP cavity polarization is almost horizontal (P-polarization) and linear (77kW ~ 89%*82.5kW)
- the angle of the polarization is roughly 2*(348-342)° = 12°
|
Accelerometer measurement, posted by Ronic Chiche at ThomX igloo about mechanics | lasers and optics | detectors and electronics
|
with Daniele, we placed the accelerometer on top of the fiber amplifier box which is inside the housing.
the goal was to try to detect a correlation with some possible accoustic noise coming from the bubbles of the water, cooling the amplifier box baseplate.
we monitored a long trend of the transmission which is perturbated when high frequency noise arises, and the accelerometer signal.
the long trend plots only the peak-peak value of a full 1 second acquisition every 2 seconds during ~2500 pts equivalent to ~5000 seconds = 1h20
on the plot, the top white signal is the transmission pk-pk and the bottom red signal is the accelerometer pk-pk.
we don't see any correlation except at 600-700 pts because Daniele entered the bunker and slightly knock on the bottom of the optical table several times.
and at 1950-2300 pts, because he opened the housing (much more noise recorded by the accelerometer) and then it close it again.
he also try to knock on the amplifier controller rack which is placed on the ground, below the table, but he didn't see any correlation with some cavity lock losses.
=> no clear conclusion.
except that the noise seen on the transmission when the housing is open is close to the "high frequency noise" we observe... could it be some accoustic noise coming from elsewhere ? |
Power shut down on September 23th, posted by Ronic Chiche at ThomX igloo about mechanics | lasers and optics | vacuum | detectors and electronics | cabling | software | utilities
|
to anticipate the power shut down on next tuesday 23/09,
this afternoon, I switched off all the equipments except the oscillator connected on the UPS. |
Power shut down on September 23th, posted by Ronic Chiche at ThomX igloo about mechanics | lasers and optics | vacuum | detectors and electronics | cabling | software | utilities
|
this morning, I restarted all the equipments of the FPC.
everything was fine after being able to get back the correct PDH phase between modulation and demodulation...I got 88kW with 33% amplifier ratio after basic alignment and CEP tuning.
the Pico1 server used for the X-ray photodiode TMD.01 doesn't want to start...
I tried with Astor. the server start to run (red => blue => green colors)
but when I start the Jive panel, it gets down again (red color)... to be solved
| Ronic Chiche wrote: |
|
to anticipate the power shut down on next tuesday 23/09,
this afternoon, I switched off all the equipments except the oscillator connected on the UPS.
|
|
Power shut down on September 23th, posted by Ronic Chiche at ThomX igloo about mechanics | lasers and optics | vacuum | detectors and electronics | cabling | software | utilities
|
the 33MHz/500MHz RF frequencies from the Ring seems different from the previous state (maybe working at an energy of 50MeV instead of 70MeV which was the previous energy ?).
so, I let the FPC at the present position and wait to discuss if it is necessary or not to move the FPC motors for the correct energy...
| Ronic Chiche wrote: |
|
this morning, I restarted all the equipments of the FPC.
everything was fine after being able to get back the correct PDH phase between modulation and demodulation...I got 88kW with 33% amplifier ratio after basic alignment and CEP tuning.
the Pico1 server used for the X-ray photodiode TMD.01 doesn't want to start...
I tried with Astor. the server start to run (red => blue => green colors)
but when I start the Jive panel, it gets down again (red color)... to be solved
| Ronic Chiche wrote: |
|
to anticipate the power shut down on next tuesday 23/09,
this afternoon, I switched off all the equipments except the oscillator connected on the UPS.
|
|
|
Power shut down on September 23th, posted by Ronic Chiche at ThomX igloo about mechanics | lasers and optics | vacuum | detectors and electronics | cabling | software | utilities
|
This morning I did a phase alignment on the PDH signal generator (Rigol DG4102)
this procedure ensure that both phases of the sine wave signals are aligned before finding the correct phase for maximizing the error signal.
so, from now on, in case of power shut down, one just need to :
1) do a phase alignment
2) put the correct phase number on the 2 channels according to the picture (if not automatically done by the generator itself).
340° on channel 1
270° on channel 2
1) and 2) can be done in any order.
| Ronic Chiche wrote: |
|
the 33MHz/500MHz RF frequencies from the Ring seems different from the previous state (maybe working at an energy of 50MeV instead of 70MeV which was the previous energy ?).
so, I let the FPC at the present position and wait to discuss if it is necessary or not to move the FPC motors for the correct energy...
| Ronic Chiche wrote: |
|
this morning, I restarted all the equipments of the FPC.
everything was fine after being able to get back the correct PDH phase between modulation and demodulation...I got 88kW with 33% amplifier ratio after basic alignment and CEP tuning.
the Pico1 server used for the X-ray photodiode TMD.01 doesn't want to start...
I tried with Astor. the server start to run (red => blue => green colors)
but when I start the Jive panel, it gets down again (red color)... to be solved
| Ronic Chiche wrote: |
|
to anticipate the power shut down on next tuesday 23/09,
this afternoon, I switched off all the equipments except the oscillator connected on the UPS.
|
|
|
|
High frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
Today, the lock of the FPC was particularly bad (maybe the worst ever seen), with a lot of high frequency noise.
the lock was impossible during several minutes !
we tried to switch off all the equipments of the machine, one by one without any effect on the lock.
at the end, we looked at the accelerometer installed inside the housing to check if there was some correlation.
and for the 1st time we clearly saw a 100% correlation beween the accelerometer signal with a noise oscillating above +/-300mV
but we didn't find the origin of this noise. |
High frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
At the end of the day, we found out what was the origin of this noise : some road renovation work with jackhammer and road roller just at the entrance of the "Igloo".
this origin was 100% correlated with a large increase of the accelerometer signal.
but we clearly saw that the FPC is much more sensitive than the accelerometer... the signal can have a small increase or just one peak and the cavity lock is lost.
then, we can make the assumption that all the "high frequency noise" which produces some lock losses could come from acoustic noise due to the road traffic or from the equipments in the bunker itself.
| Ronic Chiche wrote: |
|
Today, the lock of the FPC was particularly bad (maybe the worst ever seen), with a lot of high frequency noise.
the lock was impossible during several minutes !
we tried to switch off all the equipments of the machine, one by one without any effect on the lock.
at the end, we looked at the accelerometer installed inside the housing to check if there was some correlation.
and for the 1st time we clearly saw a 100% correlation beween the accelerometer signal with a noise oscillating above +/-300mV
but we didn't find the origin of this noise.
|
|
High frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
After this observation, we did a test with Daniele to try to correlate the road traffic in front of the Igloo with the lock losses observed on the FPC.
We didn't see any clear correlation. Cars or buses are not the direct origin of the FPC lock losses... only when some heavy load is hitting the building, we observed a clear correlation.
So, we still observed a lot of FPC lock losses when the day is windy...
One possible cause could be the large door of the igloo (~20 m²) hitting the Igloo when the day is windy.
I installed the accelerometer on the rail on the bottom of the large door to see if there is any correlation.
| Ronic Chiche wrote: |
|
At the end of the day, we found out what was the origin of this noise : some road renovation work with jackhammer and road roller just at the entrance of the "Igloo".
this origin was 100% correlated with a large increase of the accelerometer signal.
but we clearly saw that the FPC is much more sensitive than the accelerometer... the signal can have a small increase or just one peak and the cavity lock is lost.
then, we can make the assumption that all the "high frequency noise" which produces some lock losses could come from acoustic noise due to the road traffic or from the equipments in the bunker itself.
| Ronic Chiche wrote: |
|
Today, the lock of the FPC was particularly bad (maybe the worst ever seen), with a lot of high frequency noise.
the lock was impossible during several minutes !
we tried to switch off all the equipments of the machine, one by one without any effect on the lock.
at the end, we looked at the accelerometer installed inside the housing to check if there was some correlation.
and for the 1st time we clearly saw a 100% correlation beween the accelerometer signal with a noise oscillating above +/-300mV
but we didn't find the origin of this noise.
|
|
|
High frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics  
|
we did some tests with Marie to make some vibrations on the door.
below, I exhibit some pictures of the scope with the accelerometer plugged on it without or with noise...
this level of noise is not able to make the FPC loosing the lock.
even when saturating the noise signal on the accelerometer, the FPC is not loosing the lock all the time.
then, when a day is very windy, it's possible it could have some effect, but for a normal day, it's seems very doubtful that is the reason...
| Ronic Chiche wrote: |
|
After this observation, we did a test with Daniele to try to correlate the road traffic in front of the Igloo with the lock losses observed on the FPC.
We didn't see any clear correlation. Cars or buses are not the direct origin of the FPC lock losses... only when some heavy load is hitting the building, we observed a clear correlation.
So, we still observed a lot of FPC lock losses when the day is windy...
One possible cause could be the large door of the igloo (~20 m²) hitting the Igloo when the day is windy.
I installed the accelerometer on the rail on the bottom of the large door to see if there is any correlation.
| Ronic Chiche wrote: |
|
At the end of the day, we found out what was the origin of this noise : some road renovation work with jackhammer and road roller just at the entrance of the "Igloo".
this origin was 100% correlated with a large increase of the accelerometer signal.
but we clearly saw that the FPC is much more sensitive than the accelerometer... the signal can have a small increase or just one peak and the cavity lock is lost.
then, we can make the assumption that all the "high frequency noise" which produces some lock losses could come from acoustic noise due to the road traffic or from the equipments in the bunker itself.
| Ronic Chiche wrote: |
|
Today, the lock of the FPC was particularly bad (maybe the worst ever seen), with a lot of high frequency noise.
the lock was impossible during several minutes !
we tried to switch off all the equipments of the machine, one by one without any effect on the lock.
at the end, we looked at the accelerometer installed inside the housing to check if there was some correlation.
and for the 1st time we clearly saw a 100% correlation beween the accelerometer signal with a noise oscillating above +/-300mV
but we didn't find the origin of this noise.
|
|
|
|
Noise issue, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
since several days, we see that the locking is more and more unstable and the lock duration was sometime less than 20-30 seconds !
1) I tried to optimize the feedback loops
I discovered that the gain on the fast loop on the laser EOM was very low and was almost uneffective for the global stability of the lock (I can remove the cable of the HV to EOM and it does not change the stability of the signals).
when I tried to increase this loop gain in open loop for the PZT loop, I clearly saw an improvement of the transmission signal stored in the cavity.
but when I close the PZT loop, it does not help to get a better locking.
so, at the end, I cancel this loop gain (HV amplifier for the EOM is OFF !) and only the slow feedback loop on the PZT is working.
I reduced the by a factor 2 the range on the PZT loop side to reduce the noise due to the amplified Laselock (voltage divider 1/2 before the HV amp for the PZT).
and I switched off the laser motor controller (Smaract MCS).
after optmization of the PID parameters, I was able to get back the 90kW we had in the past for 33% laser amp ratio and with a very good stability :
the transmission signal is a line and the reflection signal is almost a line too.
in this condition, I seems that I can lock the laser and the FPC indefinitely.
2) issue with Smaract controller
I put back the Smaract controller only for the CEP channel => one can clearly see a bit more noise on the PDH signal but the transmssion and the reflection signal stay almost the same with a very good stability.
then, I put back the Smaract controller for both channels (CEP and laser cavity length) => the lock was very bad, even after a PID optimization despite the fact it was a "Low Vibration" (LV) Smaract MCS controller.
we tried to change the 3-channels LV controller by a standard 3-channels (not LV) spare controller => it was OK for the CEP channel but not good at all the laser cavity length channel => very noisy.
so, we kept this standard controller for the CEP channel and we took 1-channel LV spare controller for the laser cavity length channel => it was ok. no more additionnal noise and we can work without delock in piezo-scan mode.
important parameters :
I used 1ms for hold time in the Smaract controller configuration.
I used 1V/s in piezo scan mode for the speed => it does affect the noise level when the piezo is moved !
3) IP Smaract controller parameters :
I used the MCSNetworkInterfaceConfig.exe software to configure both IP addresses of the 2 controllers.
it's easy, one just has to choose the options.
the IP address for controller ruling the laser cavity length is : 192.168.1.200:5000
the IP address for controller ruling the CEP is : 192.168.1.201:5000
4) final long run test
We did a 1/2h run test to check how many lock loss one gets... see the picture.
we had some few lock losses (with the RF feedback lock ON too) but most of the time, the RF phase was not lost and the X-rays should be continued to be produced.
a test should be done tomorrow.
|
Noise issue, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
yesterday, in the same conditions, we had so many locking issues... exactly in the same way than before. :-(
so we still have "locking issues" depending on something we didn't find...
today, we removed the "1/2 voltage divider" before the PZT amplifier to get back the full dynamic.
we have a little bit more noise and we lose some power on transmission (85kW instead of 90kW) but it worked pretty well.
we had quite long X-rays runs ~ 20 minutes without any lock losses => see the picture.
but we have to keep in mind that maybe it was just a "good" day and an another day can be "bad".
we still have to understand where do these perturbations come from.
the good news is we brought back the previous setting of the FPC.
| Ronic Chiche wrote: |
|
since several days, we see that the locking is more and more unstable and the lock duration was sometime less than 20-30 seconds !
1) I tried to optimize the feedback loops
I discovered that the gain on the fast loop on the laser EOM was very low and was almost uneffective for the global stability of the lock (I can remove the cable of the HV to EOM and it does not change the stability of the signals).
when I tried to increase this loop gain in open loop for the PZT loop, I clearly saw an improvement of the transmission signal stored in the cavity.
but when I close the PZT loop, it does not help to get a better locking.
so, at the end, I cancel this loop gain (HV amplifier for the EOM is OFF !) and only the slow feedback loop on the PZT is working.
I reduced the by a factor 2 the range on the PZT loop side to reduce the noise due to the amplified Laselock (voltage divider 1/2 before the HV amp for the PZT).
and I switched off the laser motor controller (Smaract MCS).
after optmization of the PID parameters, I was able to get back the 90kW we had in the past for 33% laser amp ratio and with a very good stability :
the transmission signal is a line and the reflection signal is almost a line too.
in this condition, I seems that I can lock the laser and the FPC indefinitely.
2) issue with Smaract controller
I put back the Smaract controller only for the CEP channel => one can clearly see a bit more noise on the PDH signal but the transmssion and the reflection signal stay almost the same with a very good stability.
then, I put back the Smaract controller for both channels (CEP and laser cavity length) => the lock was very bad, even after a PID optimization despite the fact it was a "Low Vibration" (LV) Smaract MCS controller.
we tried to change the 3-channels LV controller by a standard 3-channels (not LV) spare controller => it was OK for the CEP channel but not good at all the laser cavity length channel => very noisy.
so, we kept this standard controller for the CEP channel and we took 1-channel LV spare controller for the laser cavity length channel => it was ok. no more additionnal noise and we can work without delock in piezo-scan mode.
important parameters :
I used 1ms for hold time in the Smaract controller configuration.
I used 1V/s in piezo scan mode for the speed => it does affect the noise level when the piezo is moved !
3) IP Smaract controller parameters :
I used the MCSNetworkInterfaceConfig.exe software to configure both IP addresses of the 2 controllers.
it's easy, one just has to choose the options.
the IP address for controller ruling the laser cavity length is : 192.168.1.200:5000
the IP address for controller ruling the CEP is : 192.168.1.201:5000
4) final long run test
We did a 1/2h run test to check how many lock loss one gets... see the picture.
we had some few lock losses (with the RF feedback lock ON too) but most of the time, the RF phase was not lost and the X-rays should be continued to be produced.
a test should be done tomorrow.
|
|
High frequency noise, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics
|
This morning, I installed a high-voltage probe (1:1000) on the AC-line to see if one could detect a correlation between lock losses and AC-line voltage variations.
I observed many lock losses but without any variation or correlation of the AC-line signal (scope set in "peak detect" with "envelope" arithmetics).
see attached figure.
| Ronic Chiche wrote: |
|
we did some tests with Marie to make some vibrations on the door.
below, I exhibit some pictures of the scope with the accelerometer plugged on it without or with noise...
this level of noise is not able to make the FPC loosing the lock.
even when saturating the noise signal on the accelerometer, the FPC is not loosing the lock all the time.
then, when a day is very windy, it's possible it could have some effect, but for a normal day, it's seems very doubtful that is the reason...
| Ronic Chiche wrote: |
|
After this observation, we did a test with Daniele to try to correlate the road traffic in front of the Igloo with the lock losses observed on the FPC.
We didn't see any clear correlation. Cars or buses are not the direct origin of the FPC lock losses... only when some heavy load is hitting the building, we observed a clear correlation.
So, we still observed a lot of FPC lock losses when the day is windy...
One possible cause could be the large door of the igloo (~20 m²) hitting the Igloo when the day is windy.
I installed the accelerometer on the rail on the bottom of the large door to see if there is any correlation.
| Ronic Chiche wrote: |
|
At the end of the day, we found out what was the origin of this noise : some road renovation work with jackhammer and road roller just at the entrance of the "Igloo".
this origin was 100% correlated with a large increase of the accelerometer signal.
but we clearly saw that the FPC is much more sensitive than the accelerometer... the signal can have a small increase or just one peak and the cavity lock is lost.
then, we can make the assumption that all the "high frequency noise" which produces some lock losses could come from acoustic noise due to the road traffic or from the equipments in the bunker itself.
| Ronic Chiche wrote: |
|
Today, the lock of the FPC was particularly bad (maybe the worst ever seen), with a lot of high frequency noise.
the lock was impossible during several minutes !
we tried to switch off all the equipments of the machine, one by one without any effect on the lock.
at the end, we looked at the accelerometer installed inside the housing to check if there was some correlation.
and for the 1st time we clearly saw a 100% correlation beween the accelerometer signal with a noise oscillating above +/-300mV
but we didn't find the origin of this noise.
|
|
|
|
|
|