Entry   Fast feedback loop between laser and FP cavity, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics 
    Reply   Fast feedback loop between laser and FP cavity, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics 20230907_182412.jpg20230907_182442.jpg
       Reply   Fast feedback loop between laser and FP cavity, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics 
          Reply   Fast feedback loop between laser and FP cavity, posted by Ronic Chiche at ThomX igloo about lasers and optics | detectors and electronics 
Message ID: 237     Entry time: Wed Sep 6 19:57:48 2023     Reply to this: 240
 Author: Ronic Chiche 
 Status: Fixed 
 Type: info 
 Category: lasers and optics | detectors and electronics 
 Location: ThomX igloo 
 Title: Fast feedback loop between laser and FP cavity 

today, we tried to lock the FP cavity with the Smaract motors ON (with option -LV).
we know the Smaract controllers produce some noise and the lock is very bad or impossible when the controller in ON (whatever the displacement mode is : closed or open loop, or piezo scan).
then we need to do a fast feedback loop on the EOM inside of the Onefive laser.
the problem is we cannot fill directly the error signal (~ 300mVpp) as the signal level is too low to produce some effect.

1- we tried to use the AC-coupled homemade amplifier alone but the output range is too low (+/- 3V)

2- we tried to use the M250 video amplifier for EOM with AC-coupling but the output range is still to low (+/-30V ? => to be confirmed)
but we saw an improvement in the locking.

3- we tried to combine AC-coupled homemade amplifier + M250 video amplifier for EOM with AC-coupling.
we are able to lock (~ 30% coupling) but the lock quality is very poor : we clearly see that we oscillate around the maximum of the Airy peak.
we tried several combination of the global gain (fast feedback + Laselock) using the diffuser, of the Laselock PID parameters but it seems we are not able to lock properly.

we measured a global delay of this double amplification stage of 80-90ns for the homemade amplifier and 150ns for the homemade+M250 amplifiers.
this delay is compatible with ~ 500kHz BW for the feedback => it seems it is not the reason...

we measured also the linearity of the homemade amp => there is ~ 30dB between a frequency and its 1st harmonic even at low signal...
the amp scheme is not very linear.
in comparison, the HV M250 amplifier exhibit > 80 dB of linearity !
we will try to remplace the homemade amp by a commercial FEMTO amplifier to imrpove the linearity and see if it improves also the lock.

we measured also a 100Hz AM modulation on the output signal of the homemade amp+M250.
=> we can try to work in differential (HV+ - HV-) to see if it helps to remove this modulation.