HOME SBOX THOMX MINICAV Utilities
THOMX orders THOMX installation THOMX commissioning THOMX control command
  Status of commissioning, report also here plots are reports.  Not logged in ELOG logo
Entry   Cavity Lock and Finess measurment , posted by Manar Amer at ThomX igloo about lasers and optics | detectors and electronics 6x
    Reply   Cavity Lock and Finess measurment , posted by Manar Amer at ThomX igloo about lasers and optics | detectors and electronics 
       Reply   Cavity Lock and Finess measurment , posted by Manar Amer at ThomX igloo about lasers and optics | detectors and electronics Finesse_Fit.png
          Reply   Cavity Lock and Finess measurment , posted by Manar Amer at ThomX igloo about lasers and optics | detectors and electronics apercu_03.jpgBras_manipulateur.jpg
Message ID: 127     Entry time: Fri Feb 4 10:59:56 2022     In reply to: 126
 Author: Manar Amer 
 Status: Fixed 
 Type: report 
 Category: lasers and optics | detectors and electronics 
 Location: ThomX igloo 
 Title: Cavity Lock and Finess measurment  

one thing which could reduce the Finesse is the L-shape metal piece if it is slightly inserted in the FP-cavity mode path.
understanding where this L-shape effectively is is not easy... some pictures are attached.

Manar Amer wrote:

After changing the way of injecting modulation for PDH and modulation for FSR scanning (we split the modulations to 2 different EOMs), the locking is very stable and we can measure the Finesse.

FSR = 33.34MHz

hereafter, the transmission power during an FSR scan and its fit (sweep of 300kHz of FSR in 2s)

we took 5 acquisitions which give a Finesse of :

4236
4254
4400
4045
4177

=> roughly Finesse = 4200 ! far from the 17000 previously obtained........

Manar Amer wrote:

to measure the Finesse, instead of having 2x EOM for the PDH and for the modulation sweep, I simply used 2x generators coupled with DC-blocks to a T connector (SMA) screwed directly on the EOM input.

as the transmission signal is fluctuating, it is not easy to have a good fit of the Airy peak.

if I measure the width at half of the maximum of the peak, I found roughly 10kHz instead of the awaited 2kHz... :-(

one needs a better evaluation with a more stable transmission signal and also to be sure that the L-shape metal piece (used to remove high order modes) does not introduce some losses and then reduce the Finesse...

Manar Amer wrote:

This morning, me and Ronic managed to obtain the FSR and image of the finesse which is to be analyzed.

Adding:

  • Beam Image at P4 propagation, no lens is added.
  • Image of the Finesse on Oscilloscope

more information to be included later

Will continue in the afternoon .

 

 

 