- In the last two days, Ronic and I installed new mirrors after cleaning the environment, and locked the cavity.
- We added an AOM to feedback on the high-frequency noise, but the locking condition was still not good enough. We found out that the signal generator available for this AOM has a long delay time (3 us), which may lead to low feedback bandwidth. So tomorrow we will use another AOM and signal generator to optimize the locking.
- Under this not good enough locking, we measured the finesse. Unfortunately, the finesse was measured as 15,478, which is much lower than the expected 42,000. It means that about 260ppm of additional loss was introduced. We will measure the finesse again after optimizing the locking and coupling.
By the way, attached are the delay time results for phase modulation of different signal generators:
- RIGOL DG4162: 0.7 us (best)
- SIGLENT SDG6022X: 3 us
- SIGLENT SDG7032A: 2.9 us |